Hi.
I need to repair a Tektronix 2215A. Can anyone send me the service manual?
Thanks in advance.
Martin
T
tekman
Martin said:
Hi.
Thanks in advance.
Martin
What is the problem with your 2215A?
Is the power supply not working? Tehn google, fault topic is covered
many times.
Other problem? Then let us know more about it.
hth,
Andreas
M
Martin Klein
many times.
hth,
Andreas
The display only shows a signal when pressing the "Find beam" button.
Otherwise there is no output.
I think that the horizotal deflection is not working.
I've got a Service Manual for the 2215 but the layout seems to be
different.
Any ideas?
Regards,
Martin
Jim Adney
different.
Is the beam intensity turned way down?
Jim Yanik
kl.de:
You could remove the scope case,and examine the area where the CRT neck
leads connect to the main PCB;that gives you the general area of the
horizontal output amp,and you can also check top see if a H CRT lead fell
off the CRT neck pin,which happened often enough.
The same goes for the vertical deflection.
Red and green for horizontal,brown and blue for vertical deflection leads
to the CRT.
Then you can use another scope to check for the proper sweep ramp and DC
potentials on both CRT leads.
Jim Yanik
kl.de:
neck leads connect to the main PCB;that gives you the general area of
the horizontal output amp,and you can also check top see if a H CRT
lead fell off the CRT neck pin,which happened often enough.
leads to the CRT.
DC potentials on both CRT leads.
Addendum;
For a vertical check,you can short the CRT neck pins.