HP 400F AC voltmeter

vecher

Oct 16, 2017
3
Joined
Oct 16, 2017
Messages
3
My ac voltmeter will not zero any more .., Anyone have a manual or schematic for it? Or what do I check.. It was working but lately it will not zero on any scale The lower the scale the worst..

Thanks'
Skip
 

FuZZ1L0G1C

Mar 25, 2014
366
Joined
Mar 25, 2014
Messages
366
Is it a panel meter?
Might be that needle has bent from 'pinning'.
Or plastic zero-cam may be worn or cam-pin broken.
 

dorke

Jun 20, 2015
2,342
Joined
Jun 20, 2015
Messages
2,342
Welcome to EP.
Here is the operation and service manual of the HP 400F/FL.
Are you able to get mechanical zero?
 

73's de Edd

Aug 21, 2015
3,629
Joined
Aug 21, 2015
Messages
3,629
Sir Skip . . . . .

INITIALLY . . . . . with no power applied, check to see where your meters mechanical pointer setting is having
the meter pointer being positioned, as relevant as to how close it is initially being, to scale ZERO.
Then check out the meter amp and bridge circuitry for its +- 26V's supplies and derived -6 and 9V supplies.
Also check out how well and balanced the YELLOW squared electrolytics are holding up after being 50 years old.
This was a military / aerospace microfiche schematic, so I further magged it up enough to make every small
component marking readable.
Certainly beats the service manuals piecemeal schematic sheets, if you have to travel across sections.

Re . . . .REFERENCING:

Hoolet_Plastered_400_F.jpg


73's de Edd
 

vecher

Oct 16, 2017
3
Joined
Oct 16, 2017
Messages
3
Welcome to EP.
Here is the operation and service manual of the HP 400F/FL.
Are you able to get mechanical zero?


Yes

As I go down in ranges the meter reads higher.. Does not maintain 0 Even at highest ranfe metrr is still not at 0 but pretty close.. This is with no input and terminals not shorted together..

Skip...
 

dorke

Jun 20, 2015
2,342
Joined
Jun 20, 2015
Messages
2,342
You do need to short the input terminals together, i.e. Vin=0,
the lower the input range the more important that is!
Do that with the shortest possible wire.
Still no zero?
 

(*steve*)

¡sǝpodᴉʇuɐ ǝɥʇ ɹɐǝɥd
Moderator
Jan 21, 2010
25,510
Joined
Jan 21, 2010
Messages
25,510
I recommend you check A2Q15, A2Q16, A2Q17, A2C38, A2C39, and A2C40.

These are the components commonly associated with a meter deflection with no input.

Edit: Edd has already suggested looking at those caps.
 

vecher

Oct 16, 2017
3
Joined
Oct 16, 2017
Messages
3
I recommend you check A2Q15, A2Q16, A2Q17, A2C38, A2C39, and A2C40.

These are the components commonly associated with a meter deflection with no input.

Edit: Edd has already suggested looking at those caps.


Ok I will check tomorrow.. Thanks