desertrider
- Apr 18, 2017
- 2
- Joined
- Apr 18, 2017
- Messages
- 2
Repairing a Premier Health Products PHP 2026 / 2027 Massage Chair:
I have included all service manuals attached here for this chair hoping this will help someone in the future but below is the most common problem & how to dissemble, where to get part.
Most common problem is when ALL the massage chair motors e.g. recline, Kneading, Tapping, elevation no longer work - you will notice the vibration will work.
This problem is from the donut shaped Toroid / Toroidal Transformer that powers all the motors. Attached are all the service manuals and disassembly instructions along with where to buy the part.
1. The original PHP 2026 Toroidal transformer has a single primary 110V (60Hz), dual 2x 24Vac secondary TAPS @ 16AMPs or 336VA
2. OD = Outside Diameter = 96mm (3.78”) and H = Height = 74mm (2.91”)
3. Was impossible to find a US made Toroid with this OD of 96mm – the OD must be exact to fit inside the cover or fit on the main PCB – you can find one with correct voltages but it will not fit.
4. I contacted a Toroid / Toroidal China Manufacturer (they make almost all Toroid’s in China) and they will make a custom Toroid for $50USD + $29USD=$79USD
5. The China company sells Toroid / Toroidal Transformers on eBay (been around for while) and their contact is Seller information:
a. eBay = diytalent86
b. e-mail = [email protected]
c. If seller does not exist then just look for any China Toroid / Toroidal manufacturer on eBay and contact them as they will custom make this Toroid for you - expect to wait 1 month for the part as it ships from China.
Given the US is 120Vac & the preferred output is 24Vac I decided to have the new Toroid made to meet the best input / output
Here are the SPECS in having a new custom Toroid I got made for $79.00USD:
1. (1x) Single Primary INPUT = 120Vac (US Voltage) – this can be between 110Vac to 120Vac
2. US Frequency = 60Hz
3. (2x) Dual 24Vac Secondary OUTPUT – this can output between 21-24Vac
4. Approximately 300VA to 400VA – Anywhere in this range – PHP 2026 is 336VA / 16AMP
5. Toroid OD = Outside Diameter = 96.5mm (must be 96.5mm or LESS)
6. Toroid H = Height = 73 mm – must be 73mm or less
7. There are six (6) wires coming out of the PHP 2026 Toroid:
a. Primary INPUT = Black / White is 120Vac
b. 1st SECONDARY OUTPUT = Yellow / Yellow = 24Vac (about 168VA)
c. 2nd SECONDARY OUTPUT = Blue / Blue = 24Vac (about 168VA)
d. Total SECONDARY OUTPUT MAX from original Toroid is 336VA @ 24Vac (each leg is 168VA MAX * 2 = 336VA PEAK) – this chair will not use this much the that is the rating on the original Toroid.
e. Chair will probably work find with any Toroid with power output of 200VA-400VA – I had my custom made to output PEAK 330VA.
d. This is Toroid known as a single primary with dual secondary = INPUT 110/115/120, OUTPUT 24-0-24
I have included pictures showing the measurements & label on the existing PHP 2026 Toroid Transformer.
The other attached service manuals will help you with other issues with this massage chair
desertrider
desertrider
PHP_2026-27-HOW_TO_REMOVE_PCB_boardassembly-replace.pdf385.8 KB · Views: 608
diagnostic steps for 2026 & 2027 main PCB_Reduced_Size.pdf811.3 KB · Views: 517
PHP2026_2027_technical_diagnostic_procedure_Pages_01-08.pdf889.9 KB · Views: 391
PHP2026_2027_technical_diagnostic_procedure_Pages_08-12.pdf898 KB · Views: 399
PHP2026_2027_technical_diagnostic_procedure_Pages_13-18.pdf736 KB · Views: 368