S
[email protected]
I found the igniter circuit from a short-arc xenon lamp power supply is
not reliable. most of the time, ionized plasma can be seen in lamps ,
but the lamp can't establish stable conducting current. The open
circuit voltage is ok (~120v), and current regulation is working fine
too. So I think the problem is the igniter pulse.
According to the article "Cermax lamp engineering guide" from
perkinElmer.com, peak voltage, rise time, and pulse width are all
related with triggerability. Since the power supply was working well
before, the following are the possible reasons that I'm now
considering:
1. The capacitance of the discharging (through primary winding)
capacitor decreased. (decreased trigger pulse width? or decreased peak
voltage?)
2. The dv/dt of the discharging capacitor deteriorated. (larger rise
time?)
3. the aging of the spark gap tube. the spark gap looks good, and I can
see spark inside and the sound is also a normal ping. but I don't konw
if an aging spark gap can decrease its sparkover voltage or increase
its on resistance.
any advice?
Another thing that bothers me is the estimation of the peak voltage.
The trigger transformer has 2 turns on primary and 25 turns on
secondary, and wound on a rod core (L~1.4", D~0.4", material unknown).
If I want 35KV peak voltage on secondary, how large the discharging
capacitor on the primary and how large the sparkover voltage I shall
choose? It seems to me that the larger the capacitance, the larger the
peak voltage and the pulse width on the secondary, is this true? The
power supply uses 2 6.8nF series connected ceramic disk(rated 12kV) as
the discharging capaictors. how large the sparkover voltage of the
spark gap I shall choose(assuming the capacitors can be charged up to
12kV)? Is 7.5kV too large?
Thanks!
considering:
voltage?)
time?)
its on resistance.
any advice?
Thanks!