Testing a Multimeter

Discussion in 'Electronic Basics' started by [email protected], Dec 8, 2006.

Scroll to continue with content
  1. Guest

    Hello all,

    I am looking at buying a used Fluke multimeter (either an 8060A or a
    77-III) from a local guy and I'm looking for simple tests I can do on
    each meter. Neither meter has a current calibration so I certainly
    don't expect the either to be perfect. I may send it off for a proper
    calibration after buying it, but at this point I'd just like to test
    that the meters are fully functional. I have a regulated dc supply, a
    new carbon-zinc battery and some 0.02% resistors (1,100,10K) that I
    could bring for basic circuit tests. If the dc voltage, dc current and
    resistance all checkout ok, can I assume the AC functions are ok?
    Beyond obvious signs of damage, Is there anything in particular I
    should look out for?

    Thanks,
    Jamie
     
  2. John Fields

    John Fields Guest


    ---
    Look for a transformer which you can use to check the AC voltage
    ranges with, and then, knowing that voltage and the resistances of
    your resistors, use:

    E
    I = ---
    R

    to calculate the current through the resistor(s) when they're in
    series between the transformer's output and the meter's inputs.
     
  3. Just check that all ohms and volts ranges appear to work OK. That may be
    more than enough.
     
  4. Chris

    Chris Guest

    In addition to the other advice (check AC, too -- use a transformer on
    one of the lower ranges, and also test line voltage), have the seller
    open the case (one screw on the 77-III, 4 on the 8060 AFAIR) and look
    at the circuit board. See if there's any damage or blown traces(not a
    killer in itself -- some of mine have been offered for Baptism of Fire
    and been repaired by me without problem) and make sure it looks to have
    been repaired competently. I believe there's a 1K fusible link on the
    77-III (there is one on the 77 and 77-II). It blows when you try to
    measure the resistance of the AC line. Make sure that hasn't been
    replaced with a 1K 2 watt resistor. That's a nasty thing to do --
    replace the safety component with a commonly available non-safety one,
    then sell it before she blows), but it's been done. Without a fusible
    resistor, the IC and your meter will be dead for sure on the next
    fault.

    If there is visual evidence of a repair, have the seller take out the
    assembly and look on the underside of the circuit board -- check for
    workmanship there, too. Caveat Emptor, and all that. Retest after
    putting it back together.

    If the seller won't cooperate, chances are there's something to hide.

    Good luck
    Chris
     
  5. Mr. Wizard

    Mr. Wizard Guest

    If both meters are functional you can cross calibrate to see if one of
    them is damaged. As long as the readings are withing the meter's
    specification you should be o.k. You can use your power supply and
    resistor assortment to check out the resistance, D.C. voltage and
    current ranges.

    You can do a crude check of the A.C. by checking the mains. If your
    bold enough (balls enough) you can use your resistors to exercise the
    A.C. ranges on the meter. CAUTION: IF YOU DON'T KNOW WHAT YOUR DOING,
    I WOULD WARN AGAINST DOING THIS. IT COULD BE FATAL.
     