Discussion in 'Electronic Equipment' started by [email protected], May 24, 2007.

Scroll to continue with content
  1. Guest

    Hallo,

    I tried to measure the noise of a DSP-output when no signal is applied
    to the output. I measured the "rest"-noise with an oszilloscope. The
    amplitude of the signal seems to be about 2mV. Then i wanted to
    analyse the power spectrum of the noise. I connected the DSP-output to
    a spectrum analyzer.
    Now the problem is that i get different results when changing the
    frequancy span of the spectrum analyzer. When i measure from 10Hz to
    3.2KHz, the sprectrum of the measured noise lies at about -100dBVrms.
    Increasing the span (10Hz to 25,6KHz) leads to a spectrum that lies at
    about -90dBVrms. Compared to the previous measurement it is
    approximately 10dB higher. Has anybody an explanation for this effect?
    How can i ensure a precise measurement of the noise?

    Thanks in advance

    Jens
     
  2. Fred Bartoli

    Fred Bartoli Guest

    a écrit :
    What you probably see (probably because we don't know which SA you have
    and which measurement you did exactly) is this:
    when you increase the frequency span, you also increase the SA IF
    bandwidth (most SAs have frequency span, IF and video BW coupled in
    order to make the measurement in a 'reasonable' time).

    What you want is a noise power density (nV/rtHz or uV/rtHz) and what the
    SA gives you is nV or uV. For each analyzed frequency it gives the total
    noise power that is within the IF BW around the analyzed frequency (hope
    that's clear enough). As you increase span, IF BW increases and so does
    the noise power.

    To convert to PSD you'll have to check in your manual how the noise BW
    of your SA is related to IF BW (it may not be identical, but often is),
    and then divide the noise power by the sqare root of the noise BW.

    Then the obtained figures should be identical.

    Some analyzers also have the possibility to directly do noise PSD
    measurements at some spot frequencies during the sweep. Check your
    manual for this.
     
  3. Guest


    Hello Fred,

    i hope, i got you right.
    But even if the IF (IF=input filter? antialiasing filter?) reduces the
    bandwith when a lower span is chosen, nevertheless the lower
    frequencies must be in the lower span with the same power for these
    lower frequencies are not filtered and so the power should not be less
    than with a wider span.
    If that would be the case, then the same phenomena would occure with
    much larger signals but it doesn't.
    I'm i right? Or do i get you completely wrong?

    Thanks,

    Jens
     
  4. Guest

    I have some supplements:

    Hi again,

    Now i know what you mean with IF. I don't have a heterodyne Analyser.
    I have an Analyser which is based on FFT (signal processing).
    Hope that information helps.

    Regards

    Jens
     
  5. Fred Bartoli

    Fred Bartoli Guest

    a écrit :
    In fact I was expecting that (from 25.6kHz top frequency) but this
    changes nothing.
    The bandwidth here is related to the frequency bin width, i.e.
    span/number of bins. If you have it, check your manual.
     
  6. Fred Bartoli

    Fred Bartoli Guest

    a écrit :
    IF = intermediate frequency amplifier. An heterodyne SA is nothing but a
    precision receiver.

    reduces the
    I'm not sure I understand what you mean. Well, in fact I'm sure I don't
    understand at all.

    For a given spectral noise density, the wider the filter bandwidth, the
    higher the power you'll observe at the filter output...

    ....and SA just display this power.

    To get noise spectral *density* you have to relate the observed power to
    the bandwidth.

    For a discrete signal spectrum the effect is different: widening the IF
    BW just (artificially) widens the displayed rays width.

    Have a look at agilent app note (AN150 I think) "spectrum analysis basics".
     
  7. john jardine

    john jardine Guest

    All the figures look self consistent. Would seem you have something like a
    20MHz bandwidth oscilloscope.
    The 10dB change you're seeing, emphatically implies that the spectrum
    analyser automatically increased it's measurement bandwidth by 10 times,
    when you switched from the small 3200HZ span to the much wider 25600Hz.
    You will have an option to override this 'auto bandwidth/sweep time'
    function and switch to a 'manual' filter setting.
    It a sensible world it should be described on page #1 of the operator manual
    :)
     
  8. Guest

    Hello Fred, hello John,

    thank you for your help.
    Fred: This Application Note helps a lot.
    John: Currently i have no time to care about the problem but i will
    pick your advice next week.

    Again, thank you for your help.

    Jens
     