Greetings all!
I need to consult you guys for a concern regarding lab bench instrument's and laptop grounding scheme.
I'm using the laptop (which is a class-2 device, non earthed) together withthe Pickit3 programmer/debugger. The Pickit3 is connected to the target application which is powered by an external powersupply. Connected to the application board is one oscilliscope and two waveform generators.
So, how should I earth all this to prevent anything to burn.
Between the laptop "ground" (measured on an USB connector shield)and earth I have about 80VAC when the battery charges is connected to mains.
This should be due to some Y-caps in the charger SMPS. The charger has no connection to earth so those Y-caps are probably connected between primary and secondary ground (laptop ground).
I have had the oscilloscope floating, supplied from an isolation transformer since I've worked on SMPS's lately. But now I rather have it earth connected to minimize the leakage current from the oscilloscope SMPS and it's Y-caps to my application board.
So, my question is: Is it safe for the laptop and Pickit3 to connect to the application board if I earth all the instruments. What worries me is, will that 80VAC potential cause any damage to the Pickit3 or even worse, the laptop?
Thanks for your suggestions, and have a great weekend you all!
 
P

Phil Allison

Jan 1, 1970
0
"powerampfreak"

I have had the oscilloscope floating, supplied from an isolation transformer
since I've worked on SMPS's lately.

** What a fuckwit thing to do.

Common sense says to use the ISO on the SMPS - not the damn scope.

But you have never shown any sign of sense.



.... Phil
 
M

mike

Jan 1, 1970
0
Greetings all!
I need to consult you guys for a concern regarding lab bench instrument's and laptop grounding scheme.
I'm using the laptop (which is a class-2 device, non earthed) together with the Pickit3 programmer/debugger. The Pickit3 is connected to the target application which is powered by an external powersupply. Connected to the application board is one oscilliscope and two waveform generators.
So, how should I earth all this to prevent anything to burn.
Between the laptop "ground" (measured on an USB connector shield)and earth I have about 80VAC when the battery charges is connected to mains.
This should be due to some Y-caps in the charger SMPS. The charger has no connection to earth so those Y-caps are probably connected between primary and secondary ground (laptop ground).
I have had the oscilloscope floating, supplied from an isolation transformer since I've worked on SMPS's lately. But now I rather have it earth connected to minimize the leakage current from the oscilloscope SMPS and it's Y-caps to my application board.
So, my question is: Is it safe for the laptop and Pickit3 to connect to the application board if I earth all the instruments. What worries me is, will that 80VAC potential cause any damage to the Pickit3 or even worse, the laptop?
Thanks for your suggestions, and have a great weekend you all!
Stuff breaks...but if your stuff ain't broken, it's typically OK to
connect the chassis/safety grounds together.

The problem happens when the safety ground is on the connector that
your plugging/unplugging.

I have an external hard drive with eSATA connection. A cable can connect
it directly to the motherboard. And it's FAST compared to USB.
Problem is that I am unwilling to risk my motherboard on
a glitch during plug/unpluging it. Or a static discharge that
has too much inductance to the safety ground path.

I once trashed a motherboard by plugging in a printer cable with
nothing on the other end.
As a percentage of the times I'd plugged something in, it was about zero.
But that's no consolation when your motherboard is dead.

You balance convenience with risk and take whatever chances you can
live with.

As for floating the scope, don't do it.
Use isolated probes or make a differential measurement with
probes that can stand the worst case excursion.

The use of an isolation transformer when testing any equipment
is a good idea for the same reason you use safety glasses...in case
you screw up and accidentally touch something live. It's not a license
to to connect a wire between ground and something with a lot of relative
volts on it.
 
P

Phil Allison

Jan 1, 1970
0
"mike"
As for floating the scope, don't do it.
Use isolated probes or make a differential measurement with
probes that can stand the worst case excursion.

** Or just float the circuitry with an ISO tranny.

The use of an isolation transformer when testing any equipment
is a good idea for the same reason you use safety glasses

** Not it isn't and there is no such comparison.

Routinely using an ISO tranny on *all* equipment under test /repair is
STUPID and DANGEROUS !!!

This has been thrashed out here several times now and there is NO
justification for the practice.

ONLY if you need to scope the AC line side of a device ( PSU etc) should an
ISO tranny ever be used at all.

Technicians must never get the idea that it is safe to handle ( or ground )
the live parts of a circuit - or one day it will kill them.

Plus, using an ISO hides serious fault conditions in equipment, either pre
existing or due to errors by the technician.

So, once the need for the ISO had gone, put the item back on the regular AC
supply immediately and place the ISO out of sight so there is no confusion.


.... Phil
 
P

powerampfreak

Jan 1, 1970
0
Den lördagen den 13:e april 2013 kl. 04:29:07 UTC+2 skrev Phil Allison:
"mike"








** Or just float the circuitry with an ISO tranny.









** Not it isn't and there is no such comparison.



Routinely using an ISO tranny on *all* equipment under test /repair is

STUPID and DANGEROUS !!!



This has been thrashed out here several times now and there is NO

justification for the practice.



ONLY if you need to scope the AC line side of a device ( PSU etc) shouldan

ISO tranny ever be used at all.



Technicians must never get the idea that it is safe to handle ( or ground)

the live parts of a circuit - or one day it will kill them.



Plus, using an ISO hides serious fault conditions in equipment, either pre

existing or due to errors by the technician.



So, once the need for the ISO had gone, put the item back on the regular AC

supply immediately and place the ISO out of sight so there is no confusion.





... Phil


Phil - give me a break! You talk about sense... please direct your rude comments to those who deserve it.
I must say you don't show much of a sense yourself really
Reason for ISO the scope is simply I have not a ISO tranny of 20kW which would be needed to power the SMPS.
But that's not what the topic was all about in this thred, maybe you got that wrong.
In order to connect my class-2 laptop together with the pickit3 debugger which connects to my application board which has several instruments hooked up to it. What is the best way to earth everything to minimize the risk of damage to the laptop or instruments, was the meaning of this thred.
 
N

Neon John

Jan 1, 1970
0
Phil - give me a break! You talk about sense... please direct your rude comments to those who deserve it.
I must say you don't show much of a sense yourself really

Just killfile him and he'll never annoy you again.
Reason for ISO the scope is simply I have not a ISO tranny of 20kW which would be needed to power the SMPS.
But that's not what the topic was all about in this thred, maybe you got that wrong.
In order to connect my class-2 laptop together with the pickit3 debugger which connects to my application board which has several instruments hooked up to it. What is the best way to earth everything to minimize the risk of damage to the laptop or instruments, was the meaning of this thred.

The short answer is, you don't.

The long answer is this. I'm in pretty much the same situation. The
induction heater logic boards that I design run at line voltage.
Therefore there must be galvanic isolation between it and everything.

The most important piece of kit is a USB isolator. More than once
I've forgotten or gotten in a rush and plugged a heater into line
power to do a quick firmware update. The USB isolator saved the day.

Sorry I can't give you a brand name. My partner who does all the
procurement got it from E-bay I think. It is a blob about the size of
an Aruino covered in thick heat shrink. I hi-potted it to 1500 volts
before using it.

I run my scope on an isolation transformer but that's just to protect
me from stupid. I can't make isolated measurements that way because
of the huge capacitance of the ground side vs the signal side of the
probe. For scoping, I use a differential probe similar to this

http://www.ebay.com/itm/New-Profess...457?pt=LH_DefaultDomain_0&hash=item5892516b41

If that link doesn't work just search for "differential probe". This
instrument will allow me to see things like the gate drive signal of
the high side FETs in an H-bridge.

Finally, I DO have a 20kVA isolation transformer. Actually 2 10kVA
Topaz transformers in parallel. Keep an eye out on sleazebay. They
occasionally show up and are cheap.

I have a 20kVA variac, a set of contactors and the isolation
transformers between the line and my connection to the unit under
test. The contactors are operated by a foot switch so that the UUT is
energized only when I explicitly command it to be. THAT has saved my
butt many times.

John
John DeArmond
http://www.neon-john.com
http://www.fluxeon.com
Tellico Plains, Occupied TN
See website for email address
 
P

Phil Allison

Jan 1, 1970
0
"powerampfreak"
Phil Allison:
"mike"








** Or just float the circuitry with an ISO tranny.









** Not it isn't and there is no such comparison.



Routinely using an ISO tranny on *all* equipment under test /repair is

STUPID and DANGEROUS !!!



This has been thrashed out here several times now and there is NO

justification for the practice.



ONLY if you need to scope the AC line side of a device ( PSU etc) should
an

ISO tranny ever be used at all.



Technicians must never get the idea that it is safe to handle ( or
ground )

the live parts of a circuit - or one day it will kill them.



Plus, using an ISO hides serious fault conditions in equipment, either pre

existing or due to errors by the technician.



So, once the need for the ISO had gone, put the item back on the regular
AC

supply immediately and place the ISO out of sight so there is no
confusion.


Phil - give me a break! You talk about sense... please direct your rude
comments to those who deserve it.


** For ****'s Sake Fool - have a look which post you are replying to !!

This one was NOT directed to you !!!
----------------------------------------


I must say you don't show much of a sense yourself really
Reason for ISO the scope is simply I have not a ISO tranny of 20kW which
would be needed to power the SMPS.


** That makes NO SENSE AT ALL !!

But that's not what the topic was all about in this thread, maybe you got
that wrong.


** This is now a sub thread - fuckwit !!!!

Try learning to FUCKING READ !!!

And fix your damn news reader and STOP it doubling all the line spaces.

You are an IDIOT - piss off.


..... Phil
 
P

Phil Allison

Jan 1, 1970
0
"Neon Nutcase John"
Just killfile him and he'll never annoy you again.

** YOU are a PRIZE **** aren't you ?


I run my scope on an isolation transformer but that's just to protect
me from stupid.


** No YOU are 100% stupid to have a floating scope.

I sincerely hope it kills you.
 
P

powerampfreak

Jan 1, 1970
0
I thought this was a place of engineering knowledge spreading, not a place to throw sand on each other in the kindergarden. :)
Of course my intention was not be gain enemies here...
Phil, I don't think you could judge me being an idiot or not.
Anyway, John it was interesting to hear about your high power UUT testing.
An USB isolator sounds good, but still I need to power my debugger from USB power, so I cannot use it.
Anyway, I will figure out something.
Thanks guys
 
P

Phil Allison

Jan 1, 1970
0
"powerampfreak" <[email protected]

Phil, I don't think you could judge me being an idiot or not.

** ROTFLMAO !!!!!!!!!!

Your idiotic fake name, your "hotmail" addy and

VERY SINGLE THING you post

SCREAMS:

Warning: NUT CASE ON the LOOSE !!!

**** off.
 
J

John Devereux

Jan 1, 1970
0
powerampfreak said:
I thought this was a place of engineering knowledge spreading, not a
place to throw sand on each other in the kindergarden. :)

Oh we do both here.
Of course my intention was not be gain enemies here...
Phil, I don't think you could judge me being an idiot or not.

Don't take it personally, there is indeed a lot of engineering knowledge
to be had.
 
P

powerampfreak

Jan 1, 1970
0
It's quite humoristic how angry you are Phil.
I hope it's not affecting your blood pressure too much, because I believe
you are, after all, quite a nice guy :)

Cheers
 
I thought this was a place of engineering knowledge spreading, not a place to throw sand on each other in the kindergarden. :)
Of course my intention was not be gain enemies here...
Phil, I don't think you could judge me being an idiot or not.
Anyway, John it was interesting to hear about your high power UUT testing..
An USB isolator sounds good, but still I need to power my debugger from USB power, so I cannot use it.
Anyway, I will figure out something.
Thanks guys

any real usb isolator would also isolate the power

-Lasse
 
I sort of permanently solved that problem by using Raspberry Pi and nopppas PIC programmer,
and do ssh from the laptop to it via ethernet.
Ethernet is 100% isolated, WiFi works too, even more isolated:
 http://panteltje.com/panteltje/raspberry_pi_noppp/

But it is probably OK to connect all your stuff to the same ground.
If the USB cable is OK it should ground your laptop too.
I noticed at one USB cable here that does not connect ground YMMV.

still need to power the pi so you are likely to get similar problems

the laptop you can just unplug and run on battery

-Lasse
 
J

Jasen Betts

Jan 1, 1970
0
And, last but no least, you can make Pi run from battery too, I have build
a small switcher with USB 5V out that will accept anything above about 8 V as input,
an 11 V 2200 mAh Lipo should almost last a work day, not all laptops do.

Beaglebone gives you an easy isolated power over ethenet option, rPi not so much.
 
N

Neon John

Jan 1, 1970
0
I thought this was a place of engineering knowledge spreading, not a place to throw sand on each other in the kindergarden. :)
Of course my intention was not be gain enemies here...
Phil, I don't think you could judge me being an idiot or not.

That's usenet for ya. Killfile half a dozen people and the group
becomes generally pleasant.
Anyway, John it was interesting to hear about your high power UUT testing.
An USB isolator sounds good, but still I need to power my debugger from USB power, so I cannot use it.
Anyway, I will figure out something.

My isolator has a barrel jack on it for connecting 9 volts DC to
supply the isolated side. I've never had to use it, as the isolator
passes through enough power to run my JTAG-ICE. The blob gets a
little warm so it's obvious that they're using transformer isolation
to get the power across the barrier.

If you need a specific brand, I can ask my buyer where he got it.

John
John DeArmond
http://www.neon-john.com
http://www.fluxeon.com
Tellico Plains, Occupied TN
See website for email address
 
P

Phil Allison

Jan 1, 1970
0
Neon John said:
That's usenet for ya. Killfile half a dozen people and the group
becomes generally pleasant.


** There is nothing pleasant about NGs filled with freaks and trolls.

People like you, that is.

Fucking half witted ostrich.



..... Phil
 