I'm working on a bio-potential amplifier (EMG/ECG) which is pretty
simple - just a couple of leads connected to the inverting and
non-inverting inputs of an op-amp (forget the exact part but it's
specialized differential amplifier with gain set by external resistor)
followed by a second (ordinary) op-amp for an extra X100 gain. Total
gain is adjustable from about 1,000 to 10,000.
We've had some problems with the differential amplifier blowing out
due to the use of cautery equipment (electro-surgery) or
defibrillators. A pair of 1N914 diodes wired across the inputs didn't
help.
When the differential amplifier blows, some sort of leakage current
back-feeds from the amp, through the leads, and into the heart
(causing fibrillation). This is not a human/clinical application.
So we try putting capacitors (0.1 uf) in series with each of the two
input leads. The theory is that if the amp blows it won't be able to
back-feed any current through the leads.
But I've noticed that at the low frequencies of ECG signals (the QRS
complex resembles a single cycle of 10 to 15 hz sinusoid) that the
output of the amplifier actually LEADS the input by about 8 ms.
While feeding in burst sine waves of higher frequencies, the effect
disappears at higher frequencies (say, 50 to 100 hz).
I was wondering what sort of mathematics can explain this.
simple - just a couple of leads connected to the inverting and
non-inverting inputs of an op-amp (forget the exact part but it's
specialized differential amplifier with gain set by external resistor)
followed by a second (ordinary) op-amp for an extra X100 gain. Total
gain is adjustable from about 1,000 to 10,000.
We've had some problems with the differential amplifier blowing out
due to the use of cautery equipment (electro-surgery) or
defibrillators. A pair of 1N914 diodes wired across the inputs didn't
help.
When the differential amplifier blows, some sort of leakage current
back-feeds from the amp, through the leads, and into the heart
(causing fibrillation). This is not a human/clinical application.
So we try putting capacitors (0.1 uf) in series with each of the two
input leads. The theory is that if the amp blows it won't be able to
back-feed any current through the leads.
But I've noticed that at the low frequencies of ECG signals (the QRS
complex resembles a single cycle of 10 to 15 hz sinusoid) that the
output of the amplifier actually LEADS the input by about 8 ms.
While feeding in burst sine waves of higher frequencies, the effect
disappears at higher frequencies (say, 50 to 100 hz).
I was wondering what sort of mathematics can explain this.