Steve said:
No clue. The original is history, the customer plugged the unit into
230V. No way to see any case markings or even physical dimentions of
the original.
I don't know exactly what technology that is, but it's function is a surge
suppressor. The unit will run if that part is removed. Whether anything
else has been damaged, only way to know is to try it. I'd suggest checking
D1, D2, C1, and R3. If they are good (or after replacing whatever isn't!),
try it without D3.
130 V may be a bit close to 125 VAC, maybe 140 or 150 V? Replace with
modern device and add a fuse while you're at it.
With the fuse, even
if D3 isn't replaced with anything, a surge or plugging into 230 VAC shouldn't
result in damage beyond D1 in any case.