Fluke 8600a meter repair

Discussion in 'Electronic Repair' started by [email protected], Jul 26, 2005.

Scroll to continue with content
  1. Guest

    Hi,

    I have 7 8600a here with the same problem, several mV Offset in the
    200mV AC range which leads to severe offset in all mA AC ranges (3
    digits). DC ranges are fine.

    I have more details if there's someone out there with experience with
    the 8600a. Just can't seem to find the fault. On the output of the AC
    converter i see a difference between output, and AZ out, with shorted
    input, which causes the offset. But more details if there's actually
    someone who knows about the 8600a.

    thanks

    ST
     
  2. Guest

    Turns out my stupid desk lamp fouls the measurement. Is a fluorescent
    with electronic ballast, should've known better.

    Still some offset, but it appears within tolerance. Bad shielding there
    i'd say.

    Still interested in discussing the problem though, but much of the
    mystery is solved.

    ST
     