Discussion in 'Troubleshooting and Repair' started by Moudy, Oct 8, 2016.

Scroll to continue with content
  1. Moudy

    Moudy

    2
    0
    Oct 8, 2016
    my DT9205A voltmeter gives faulty read of ac , when I try to test 220v ,50Hz " Home Current" it give 100v and it keep droping to 5v , and the rest of the voltmeter functions are OK , works fine , just the AC test is faulty, so there is anyway to fix it or replace the damaged part ?


    Thanks
     
  2. Minder

    Minder

    2,957
    625
    Apr 24, 2015
    With my Fluke I get Flaky readings if the battery is low, this is common to these meter, could be the same case with yours.
    My errors before the battery low ind comes on.
    M.
     
    Moudy and davenn like this.
  3. Moudy

    Moudy

    2
    0
    Oct 8, 2016
    Thanks , you were right , it was the battery ,but I didn't think of it because the voltmeter has low battery indicator which didn't work

    Again thank you for your fast reply , I appreciate it .
     