Gooday to all out there!

This is my first post on your forum so be gentle with me.

I have a Tektronix 2445B scope that shows "Test 05 Fail 44" on startup
"Press A/B Trig To Clear" which when pressed all seems well.

What is this error message all about?
Will it prove fatal and can it be fixed if so how?

Thanks for reading hope you can help

Regards Bazzatron
Western Australia.












'
 
Reply With Quote
 
 
 
 
Senior Member
Join Date: Jan 2010
Posts: 1,026
 
      08-12-2010, 06:07 PM
The "press A/B trig to Clear" automatically comes up when any one of the power-up test fails, ....that's normal, it's working correctly.
When you power-up, scope runs it's own internal test routine, one test failed, giving you the above notification.
Routine Test #05 is a test on the main board.
The error code is telling you the Positive Level is Too Positive.
(I'm a little busy here right now, when I've got some time, I'll check into the solution for this error, and post another reply with additional info.)
 
Reply With Quote
 
Senior Member
Join Date: Jan 2010
Posts: 1,026
 
      08-12-2010, 06:32 PM
Ok, first I'll tell you probably what you want to know:
Your error code is not fatal, and you can keep using the scope just the way it is with no problems.
Now I'll tell you what's going on with it, in case you're interested, but I STRONGLY suggest you don't do anything about it yourself. This is just for your information only.
The test that generated the error code detected a high peak on the LEVEL of your line trigger signal. There isn't any potentiometer to adjust THAT, it's all done in the electronics. What a repair shop would do, is run the 'self-calibration' for the instrument, (DON'T do this yourself). The repairman would remove a jumper inside the instrument to enable the self calibration, and adjust horizonal, vertical, and trace pots to set the scope levels. But you ABSOLUTELY cannot do that without a lab standard calibrated DC voltage source. If you adjusted anything without an exact DC cal standard, you'd screw the scope alignment royally, and wouldn't know where you were at with it. I suppose you didn't NEED to know that, but since you're on this site, asking questions, I thought you might LIKE to know.
You're fine with the scope the way it is, it's just telling you the trigger level is a little too positive (but don't touch it yourself).
 
Reply With Quote
 
Super Moderator
(*steve*)'s Avatar
Join Date: Jan 2010
Location: Mid way between Beijing and the the Ronne Ice Shelf (mind your projection) ...and don't call me Sir!
Posts: 7,331
Send a message via Skype™ to (*steve*)
 
      08-13-2010, 12:45 AM
Quote:
Originally Posted by Bazzatron View Post
Western Australia.
'
G'day from a sandgroper.
 
Reply With Quote
 
Junior Member
Join Date: Aug 2010
Posts: 5
 
      08-13-2010, 05:44 AM
Gooday all,

Thankyou for your prompt reply to my questions regarding the startup fault displays.

Whew! You have certainly put my mind to rest I had a suspicion that nasty things
were about to happen.

I am well aware that this scope has a few years on its clock and could and probably will develop faults, to this end i have downloaded service and operators manuals but they are
not always clear as to what if any procedures need to be done.
Is there any way to read from the scope its time-on-in-use data?
It would be interesting to know how many hours its done etc.

Well rest assured i will be a good boy and leave well enough alone in this instance!


Thanks again

Barry.
 
Reply With Quote
 
Senior Member
Join Date: Jan 2010
Posts: 1,026
 
      08-13-2010, 12:57 PM
I don't know of any way to find out actual hours on the scope.
Mfgr says it was built for 2,000 hours of no-problem use (just for your info).
As always, recommended calibration from the mfgr is once a year.
But Calibration is always expensive. And your problem is not a serious one.
In your case, I'd run it until you start having other problems, then consider calibration.
Just my opinion.
 
Reply With Quote
 
Junior Member
Join Date: Aug 2010
Posts: 5
 
      08-13-2010, 02:38 PM
Thanks for your efforts regarding the info for finding Hours of use.

I will read through the service and operators manuals and if i find
a reference to reading hours of use i will post the results here.

Thanks again everyone
Regards from Down Under
Bazzatron
 
Reply With Quote
 
Junior Member
Join Date: Aug 2010
Posts: 5
 
      08-13-2010, 03:55 PM
Sucess!!
I have found the procedure to show the hours of use and power cycle count.

Using the 2445B/2455B/2465B Oscilloscope And Options Operators Manual
070-6860-00

Appendix A

Page177

Extended Functions With Diagnostic Exercisers

Exer 05
Display Operating Time And Power Cycle Count
Is the Exer to use as per the procedure on page 177
And page 178 shows how Exer 05 displays the data when it is run.


Whew! these tektronix people sure know how to write a tech manual
all the info is there IF you can find it!

Well guys hope that info will be of value to others wanting to know how many hours run and power on cycles their 2445B scope has done and to keep track of times between calibrations.

Regards
And Thanks again for all your valued input to my questions

Bazzatron
WA
 

Last edited by Bazzatron; 08-13-2010 at 03:57 PM..
Reply With Quote
 
Super Moderator
(*steve*)'s Avatar
Join Date: Jan 2010
Location: Mid way between Beijing and the the Ronne Ice Shelf (mind your projection) ...and don't call me Sir!
Posts: 7,331
Send a message via Skype™ to (*steve*)
 
      08-14-2010, 12:12 AM
So, come on, how many operating hours does it have on it?
 
Reply With Quote
 
Junior Member
Join Date: Aug 2010
Posts: 5
 
      08-24-2010, 01:54 PM
Sorry about the delay in reply

I ran the routine and got 29,276 total hours on and 3,152 on-off cycles.

works out to approx 3.36 years of use since built

assuming build date was 1989 or 1990 thats not bad

3.36 years of use in 20 years

Am having further issues now with inability to reduce intensity
will investigate and advise when i find the problem
i suspect bad caps etc
 
Reply With Quote
 
 
 
Reply

Thread Tools
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are Off


Similar Threads
Thread Thread Starter Forum Replies Last Post
TEKTRONIX 2465 fault Stevescope Electronic Equipment 5 06-08-2010 08:56 PM
Tektronix 2445B Scope Problem , Help needed (Video Included) NetSpid3r@gmail.com Electronic Repair 17 02-25-2010 06:29 AM
Tektronix 2445b with problem Bestmimo Electronic Repair 9 04-05-2009 05:23 AM
Re: tektronix 2215 fault T Misc Electronics 0 05-05-2008 03:11 AM
Help with Tektronix TDS 460 scope acquisition fault please Smeghead Electronic Repair 12 02-22-2008 06:41 AM



1 2 3 4 5 6 7 8 9 10 11 12 13 14 15 16 17 18 19 20 21 22 23 24 25 26 27 28 29 30 31 32 33 34 35 36 37 38 39 40 41 42 43 44 45 46 47 48 49 50 51 52 53 54 55 56 57 58 59 60 61 62 63 64 65 66 67 68 69 70 71 72 73 74 75 76 77 78 79 80 81 82 83 84 85 86 87 88 89 90 91 92 93 94 95 96 97 98