How to perform 200kHz-5MHz sweep while maintaining 200mArms?

 
 
Michael
Guest
Posts: n/a
 
      09-22-2004, 09:43 PM
I am performing a test to measure the impedance through electrode pads
placed on opposite parts of the human body. The protocol requires that
the measurement be made by maintaining an AC sine current of 200mArms,
sweeping the frequencies from 200kHz to 5MHz, and measuring the Vrms
between one electrode and a reference point on the body. The impedance
is to be calculated by dividing this voltage by the known 200mArms
current.

How do I sweep frequencies while maintaining a constant 200mArms
current using a HP 33120A (waveform generator w/ 0-15MHz freq and
0-10V peak to peak) or a HP 4192A (impedance analyzer w/ 0-15MHz freq
and -1V-+1V peak to peak)? Can I construct a circuit external to one
of these units and accomplish the 200mArms sweep? If not, what
lowest-cost set-up could?
 
Reply With Quote
 
 
 
 
Walter Harley
Guest
Posts: n/a
 
      09-23-2004, 12:21 AM
"Michael" <(E-Mail Removed)> wrote in message
news:(E-Mail Removed) m...
>I am performing a test to measure the impedance through electrode pads
> placed on opposite parts of the human body. The protocol requires that
> the measurement be made by maintaining an AC sine current of 200mArms,
> sweeping the frequencies from 200kHz to 5MHz, and measuring the Vrms
> between one electrode and a reference point on the body. The impedance
> is to be calculated by dividing this voltage by the known 200mArms
> current.
>
> How do I sweep frequencies while maintaining a constant 200mArms
> current using a HP 33120A (waveform generator w/ 0-15MHz freq and
> 0-10V peak to peak) or a HP 4192A (impedance analyzer w/ 0-15MHz freq
> and -1V-+1V peak to peak)? Can I construct a circuit external to one
> of these units and accomplish the 200mArms sweep? If not, what
> lowest-cost set-up could?



See quasi-answer on sci.electronics.components. And, please learn to
cross-post, rather than posting multiple times, so that your messages will
be linked across the newsgroups. Thanks!


 
Reply With Quote
 
Jim Yanik
Guest
Posts: n/a
 
      09-23-2004, 01:27 AM
(E-Mail Removed) (Michael) wrote in
news:(E-Mail Removed) m:

> I am performing a test to measure the impedance through electrode pads
> placed on opposite parts of the human body. The protocol requires that
> the measurement be made by maintaining an AC sine current of 200mArms,
> sweeping the frequencies from 200kHz to 5MHz, and measuring the Vrms
> between one electrode and a reference point on the body. The impedance
> is to be calculated by dividing this voltage by the known 200mArms
> current.
>
> How do I sweep frequencies while maintaining a constant 200mArms
> current using a HP 33120A (waveform generator w/ 0-15MHz freq and
> 0-10V peak to peak) or a HP 4192A (impedance analyzer w/ 0-15MHz freq
> and -1V-+1V peak to peak)? Can I construct a circuit external to one
> of these units and accomplish the 200mArms sweep? If not, what
> lowest-cost set-up could?


I don't think your generator(s) will output 200ma into a load.
BTW,flatness of your generator probably depends on a 50 ohm load Z.

--
Jim Yanik
jyanik-at-kua.net
 
Reply With Quote
 
Ed Price
Guest
Posts: n/a
 
      09-23-2004, 11:11 AM

"Michael" <(E-Mail Removed)> wrote in message
news:(E-Mail Removed) m...
>I am performing a test to measure the impedance through electrode pads
> placed on opposite parts of the human body. The protocol requires that
> the measurement be made by maintaining an AC sine current of 200mArms,
> sweeping the frequencies from 200kHz to 5MHz, and measuring the Vrms
> between one electrode and a reference point on the body. The impedance
> is to be calculated by dividing this voltage by the known 200mArms
> current.
>
> How do I sweep frequencies while maintaining a constant 200mArms
> current using a HP 33120A (waveform generator w/ 0-15MHz freq and
> 0-10V peak to peak) or a HP 4192A (impedance analyzer w/ 0-15MHz freq
> and -1V-+1V peak to peak)? Can I construct a circuit external to one
> of these units and accomplish the 200mArms sweep? If not, what
> lowest-cost set-up could?



Before you get too deep figuring out "how" you do this, have you determined
"if" you may do this? I sure don't want anybody to put 200 milliamps through
me! Assuming a typical body resistance of about 1500 Ohms, doesn't that
imply about 60 Watts? True, I don't know what happens to body impedance as
you get into the low HF range, but maybe you should review the literature on
tissue heating one more time. Perhaps all your volunteers are already
cadavers. Or maybe you Berkeley guys don't know milliamps from microamps.

Ed
wb6wsn

 
Reply With Quote
 
Mark
Guest
Posts: n/a
 
      09-23-2004, 04:08 PM

"Ed Price" <(E-Mail Removed)> wrote in message
news:Lby4d.56609$9Y5.39735@fed1read02...
>
> "Michael" <(E-Mail Removed)> wrote in message
> news:(E-Mail Removed) m...
> >I am performing a test to measure the impedance through electrode pads
> > placed on opposite parts of the human body. The protocol requires that
> > the measurement be made by maintaining an AC sine current of 200mArms,
> > sweeping the frequencies from 200kHz to 5MHz, and measuring the Vrms
> > between one electrode and a reference point on the body. The impedance
> > is to be calculated by dividing this voltage by the known 200mArms
> > current.
> >
> > How do I sweep frequencies while maintaining a constant 200mArms
> > current using a HP 33120A (waveform generator w/ 0-15MHz freq and
> > 0-10V peak to peak) or a HP 4192A (impedance analyzer w/ 0-15MHz freq
> > and -1V-+1V peak to peak)? Can I construct a circuit external to one
> > of these units and accomplish the 200mArms sweep? If not, what
> > lowest-cost set-up could?

>
>
> Before you get too deep figuring out "how" you do this, have you

determined
> "if" you may do this? I sure don't want anybody to put 200 milliamps

through
> me! Assuming a typical body resistance of about 1500 Ohms, doesn't that
> imply about 60 Watts? True, I don't know what happens to body impedance as
> you get into the low HF range, but maybe you should review the literature

on
> tissue heating one more time. Perhaps all your volunteers are already
> cadavers. Or maybe you Berkeley guys don't know milliamps from microamps.
>
> Ed
> wb6wsn



oh I hope it is microamps.... I've heard that 30millamps can kill!



 
Reply With Quote
 
Walter Harley
Guest
Posts: n/a
 
      09-23-2004, 06:15 PM
"Mark" <bigfathhairybiker@MAILhot(swapemround).com> wrote in message
news:4152f420$0$20246$(E-Mail Removed). net...
> oh I hope it is microamps.... I've heard that 30millamps can kill!


Less than that can kill; but it has to actually get to the relevant muscles
(heart, diaphragm).

In this case, the *intent* of the OP was to use frequencies of 200kHz and
above. At those frequencies, the energy travels mainly on the surface of
the conductor, that is, on the surface of the body. Apparently, there is
not even any sensation of being shocked, even at fairly high currents.

If you do a bit of Googling for "high frequency current human body" (without
the quotes) you can learn more about this.

The key point, though, is that word "intent". As we all know, what actually
happens when you hit the switch on a piece of equipment is not always
exactly what we intended to happen. There is altogether too much risk that
an equipment fault or an operator error would result in substantial current
being delivered at lower, lethal, frequencies.


 
Reply With Quote
 
 
 
Reply

Thread Tools
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are Off


Similar Threads
Thread Thread Starter Forum Replies Last Post
maintaining 12 V aqe911@hotmail.com Electronic Basics 2 05-30-2008 04:38 PM
maintaining constant temperature? leon.park@gmail.com Electronic Basics 12 11-17-2005 11:05 PM
Who is maintaining the RS-274X standard ? Adam S. CAD 3 01-07-2005 09:30 PM
How to perform 200kHz-5MHz sweep while maintaining 200mArms? Michael Electronic Basics 10 09-23-2004 07:32 PM
How to perform 200kHz-5MHz sweep while maintaining 200mArms? Michael Electronic Components 2 09-23-2004 12:36 AM



1 2 3 4 5 6 7 8 9 10 11 12 13 14 15 16 17 18 19 20 21 22 23 24 25 26 27 28 29 30 31 32 33 34 35 36 37 38 39 40 41 42 43 44 45 46 47 48 49 50 51 52 53 54 55 56 57 58 59 60 61 62 63 64 65 66 67 68 69 70 71 72 73 74 75 76 77 78 79 80 81 82 83 84 85 86 87 88 89 90 91 92 93